Recognising ion suppression
One method is to put a very low concentration of the metabolite of interest in the solvents you are using to run the chromatography. It must be a very low concentration, because most such metabolites are not volatile, and it is bad to put a lot of non-volatile material into a machine that relies on evaporating everything. With the metabolite in the solvent, inject a sample and carry out a normal run.
If you have a syringe pump then you can fit a T-piece between the column and the detector and gently infuse metabolite from the syringe pump into a run, post column.
You should now have a very high base-line, caused by a continuous start-to-finish presence of the metabolite of interest. If a chemical is eluting that causes ion suppression, it will suppress the metabolite in the baseline, and you will see a negative peak.
The example below was not an intentional suppression experiment. It was a (horrible) run where the user had accidentally included something that ionised in one of their buffers, while running a gradient. Therefore they had a gradually increasing base-line. But at about 29 minutes something eluted from the column that did not itself ionise, but suppressed the background ionisation. This caused the negative peak...